제품특징 | 제품사양 | 자료실 |
- 2 universal analog input channels
Voltage, current, resistance, potentiometer, Pt100, Pt1000, thermocouples, measuring bridges, IEPE sensors
- Fast high accuracy digitalization
24 bit ADC, 100 kHz sample rate per channel
- 1 digital in or output per channel
Input: state, tare, memory reset
Output: state alarm, threshold
- Signal conditioning
16 virtual channels, linearization, digital filter, average, scaling, min/max storage, RMS, arithmetic, alarm
- Galvanic isolation
Of I/O-signals, power supply and interface
Isolation voltage 500 VDC
제품특징 | 제품사양 | 자료실 |
Analog Inputs
Channel Number Accuracy Linearity error Repeatability Isolation voltage Measurement Type |
2 0.01 % typical 0.02 % in controlled environment 0.05 % in industrial area 0.01 % of the final value typical 0.003 % typical (within 24 h) 500 VDC channel to channel to power supply to isolation voltage voltage, current, resistance,
potentiometer, Pt100, Pt1000, thermocouples, full and half bridges,
IEPE-sensors |
Analog/Digital-Conversion
Resolution Sample rate Conversion method Anti-aliasing filter Digital filter Averaging |
24 bit 100 kHz (measurement thermocouple 10 Hz) Sigma-Delta (group delay time 380 μs) 20 kHz, 3rd order IIR, low pass, high pass, band pass, 4th
order, 1 Hz up to 10 kHz in steps 1, 2, 5 configurable or automated according the
selected data rate |
Digital In/Outputs
Number Response time Input Input voltage Input current Lower / upper threshold Output Contact Load |
2 (1 digital I/O per channel) 0.2 ms state, tare, reset max. 30 VDC max. 0.5 mA <2.0 V (low) / >10 V (high) state, alarm open drain p-channel MOSFET 30 VDC / 100 mA (ohmic load) |
Power Supply
Power supply Power consumption Influence of the voltage |
10 up to 30 VDC, overvoltage and overload
protection approx. 2 W <0.001 %/V |
Environmental
Operating temperature Storage temperature Relative humidity |
-20°C up to +60°C -40°C up to +85°C 5 % up to 95 % at 50°C, non condensing |
Dimension
Front plate (W x H) Depth |
(30 x 128) mm 118 mm |