제품특징 | 제품사양 | 자료실 |
- 4 galvanically isolated input channels
for non-insulated thermocouples at high potential
- Insulation voltage 1200 VDC
short-term 5000 VDC
- Cold junction compensation
- Dynamic linearization
Optimum positioning of interpolation points in selected range, types B, E, J, K, L, N, R, S, T, U
- High accuracy digitalization
24 bit ADC, 10 kHz sample rate per channel
- Signal conditioning
digital filter, average, scaling, min/max storage, arithmetic, alarm
제품특징 | 제품사양 | 자료실 |
Analog Inputs
Channel Number Accuracy Linearity error Repeatability Isolation voltage Measurement Type |
4 0.01 % typical 0.025 % in controlled environment 0.05 % in industrial area 0.01 % of the final value typical 0.003 % typical (within 24 h) 1200 VDC permanent, channel to channel to power supply thermocouple on high voltage potential (not isolated), isolation voltage 1200 VDC permanent |
Analog/Digital-Conversion
Resolution Sample rate Conversion method Anti-aliasing filter Digital filter Averaging |
24 bit 20 kHz each channel Sigma-Delta (group delay time 600 μs) 1000 Hz, 2nd order IIR, low pass, high pass, band pass, 4th
order, 1 Hz up to 10 kHz in steps 1, 2, 5 configurable or automated according the
selected data rate |
Power Supply
Power supply Power consumption Influence of the voltage |
10 up to 30 VDC, overvoltage and overload
protection approx. 2 W <0.001 %/V |
Environmental
Operating temperature Storage temperature Relative humidity |
-20°C up to +60°C -40°C up to +85°C 5 % up to 95 % at 50°C, non condensing |
Mechanical
Case Dimensions (W x H x D) Weight Mounting |
Aluminum and ABS (27 x 120 x 105) mm approx. 200 g DIN
EN-rail |